I have to control (part of an own application) a dataman S4 over RS232-interface. Can someone tell me the exactly syntax (specifically the address-range!) which i have to send to RS232-interface to load a device. When sending "LO 00000,000FF,00000" to RS232, the dataman S4 displays following message "Address 000xx". After this no ">"-prompt will be displayed.
I also tried to send "LO00000000FF<cr>", with the same result!
Whats wrong?
thanks
Pascal Soland
Controlling S4 over RS232, Error-message "Address 000xx"...
-
- Posts: 931
- Joined: 10 Nov 2011 09:51
Re: Controlling S4 over RS232, Error-message "Address 000xx"
Dear Guest,
The load command should be in the following format:
LOAD (START)-(END)=(DEST)
LOAD copies the contents of the PROM in the ZIF into USER-RAM. The area copied and the destination is defined by the START and END addresses.
I hope this helps.
Best regards,
The load command should be in the following format:
LOAD (START)-(END)=(DEST)
LOAD copies the contents of the PROM in the ZIF into USER-RAM. The area copied and the destination is defined by the START and END addresses.
I hope this helps.
Best regards,